| comentario del autor | Vie Oct 16, 2009 11:48 am | |
|
|
||
| sin valorar | Vie Oct 16, 2009 12:21 pm | |
|
AUTO AND DEALER CALIBRATION FOR HS SERIES DEALER CALIBRATION PROCEDURE 1. Turn scale OFF 2. Press and hold MODE and turn scale on 3. Scale displays “F.1” 4. Press HOLD 5. Scale displays offset value 6. Press ON/ZERO 7. Press MODE 8. Press ON/ZERO 9. Press MODE 10.Press ON/ZERO. Scale displays a arrow on the at the right hand of display. Make sure figure displayed is =0 or 1. If not, press ON/ZERO again. 11.Load weight for calibration as listed above 12.When value displayed is stable press HOLD 13.Wait till scale starts countdown 14.Calibration complete and scale is ready for operation |
||
| comentario del autor | Vie Oct 16, 2009 4:16 pm | |
|
|
||
| sin valorar | Sab Oct 17, 2009 6:53 am | |
|
En sí, cual es la falla de la balanza ( ? ) Precísamela. Puedes realizar los ajustes de Zero y Span, mediante Trimmers. Tener a la mano patrones con peso exacto. |
||
| sin valorar | Sab Oct 17, 2009 9:16 pm | |
|
debajo de la tarjeta hay un trimpot que es para calibrar (span fino) tambien puedes puedes acceder por debajo dandole vuelta a la balanza y si el error excede el rango del trimpot utiliza los jumper (span grueso). Saludos. |
||