Access to the calibration adjustment is prevented by setting a switch (located below a small access cover on the right side of the indicator) to the locked position. This is sealed by placing a destructible adhesive label over the access cover, and also over an access screw in the battery compartment of the indicator . It is important to ensure that the switch is in the locked position. This may be checked (without removing the sealing) by the following procedure: • Turn the instrument off and then on using the switch on the left side of the indicator. • Whilst the power on test sequence is displayed, press and hold the ZERO key until the display shows “02001”. • Release the ZERO key. • “F0” is displayed – press the enter ( <--l ) key. • If the calibration switch is in the locked position the display will show “02001”. Switch the instrument off and then on to return to normal operating mode. • If the calibration is not in the locked position the display will show “Zero” (the instrument is not correctly sealed). Switch the instrument off and then on to return to normal operating mode. adj.inst.calib. http://www.yoreparo.com/foros/download.php?id=24727 |
||